Heraeus multifuge 3S Error 26

Hello,

I have lost the booklet that came with our multifuge, so was wondering if you could help me out by telling me what Error 26 means please?

Thank you very much,

Erin


Report Post
Wed Feb 27 2013Reply from MedWrenchManager
MedWrenchManager

You should try contacting Ozark Biomedical, they may be able to help


Report Post
Wed Feb 27 2013Reply from ErinFirmin
avatar placemark

Thank you :)


Report Post
Wed Feb 27 2013Reply from C. Mash
C. Mash

Error 26 states "Time out - latch close" which usually is the latch motor sensor pcb. The best solution for this would be removing the hub attached to the latch motor shaft and cleaning the magnets on the back side of it. If that does not correct the issue, the sensor pcb may need replacing. Thanks again for using Med-Wrench.


Ozark Biomedical
Senior Technical Specialist
800-457-7576
Report Post
Wed Feb 27 2013Reply from ErinFirmin
avatar placemark

Thank you so much for your reply, I will definitely try that! :)