How hard is the PM and are there alternatives??

Thank you


   RE: Autotransfusion Unit Report Post
Mon Jul 29 2013Reply from Cindy
avatar placemark

Mike,

The Pm is not hard but it does require you verify the calibration of the Bowl optics sensor, which requirs an optics aiming fixture ETD#5398. I don't recall the cost of the aiming fixture but it does require annual certification too. The rest of the pm can be done with standard test equipment, DVM, DPM, calibrated wieghts and a disposbale set for the cell saver. We had enough units where Haemonetics comped our school. The class was very informative. Their tech support is very helpful too.


Report Post
Fri Jun 27 2014Reply from BioMedovation
avatar placemark

Soaring Hearts is the only third party alternative that I know of that can work on and PM perfusion equipment. You can contact them at 1-855-438-7744


Report Post