ADAC genesys analog gantry will not boot

Analog dual head gantry will not boot. Power supplies all good but mpu and relays keep cycling/resetting themselves. Don't have a good diagram of analog gantry.

Thinks possible bad Mpu or safety circuit is not kicking in properly. E stops all appear to be ok as well. 24v to Mpu seems solid but need to scope.

Any help is appreciated!


   RE: Nuclear Medicine-Gamma Camera Report Post
Wed Sep 18 2013Reply from Ralph!
avatar placemark

Try checking ripple on the power supplies.Should be less then 100v.That can cause crazy things to happen to the unit.


Report Post
Mon Sep 23 2013Reply from lee0857
avatar placemark

typically what happens are a couple of things:

1: The battery backup (the black rectangular thing) on the gantry mpu is dead...

2: If it is dead (you have to remove mpu to check on the back of the board) you can use a phone battery (3.5 vdc) in its place. ie red/black wires and solder to the mpu where the battery pack used to be

3: once you replace the battery you must go into dos and run gantcal.exe and restore the cal and limit files to the gantry mpu

4: once you complete this you must recal the gantry and run gantcal again and save the results

5: it is assumed that you know anything about what im saying and you have a basic understanding of how to move around in the file structure.

There you have it.