Error service 24AB0

Error service 24AB0 is holding the caset inside.

 


   RE: Computed Radiography Report Post
Thu Jan 16 2014Reply from Ali-ahmad
avatar placemark


 Handling Control Module:  Vacuum error postscan.
Got no vacuum  while taking IP from scanner.
 Probably IP Jam

open doors and put IP into cassette.


Report Post
Mon Jan 25 2016Reply from Imaging Service Engineer I
avatar placemark

This error is caused by the pickup suction cups on the output side of the laser not being tilted at the right angle and cannot mate to the plate.  Suction cup(s) need to be adjusted for tilt angle.