When switch on the machine it doesn't make the self test and when start the case it gives an alarm (Alternate O2 Screen), so should I replace the mixer or just one valve?

 


   RE: Anesthesia Gas Machine Report Post
Fri Jan 10 2014Reply from Aneserv Medical Inc.
Aneserv Medical Inc.

You have to go through the troubleshooting procedure in the service manual. Try zeroing the pressure transducers or check the alt O2 switch...could be simple and not the mixer. The service menu has extensive tests/info for the machine that should point to the problem. 


Report Post
Sat Jan 11 2014Reply from Ihsan
avatar placemark

My analysis that there is something wrong in the mixer that's way the machine asks to open the alt O2,is this could be right?