Mon Feb 10 2014

Model XEO giving 322 and 367 errors.

What service is required to clear 322 and 367 errors?


   RE: Aesthetic Laser Report Post
Mon Feb 10 2014Reply from MedWrenchManager
MedWrenchManager

Who is the manufacturer?

 


Report Post
Mon Feb 10 2014Reply from BetterBiomed
avatar placemark

Cutera (XEO Cosmetic Laser)


Report Post
Wed May 14 2014Reply from CRC Biomedical Services
CRC Biomedical Services

322 is a water flow fault. 367 usually accompanies other faults- not 322, however. You may have at least one more fault.


Mickey A. Couvertier, CBET, CLRT
President & Senior Technician
CRC Biomedical Services
http://www.crcbiomedical.com