what are the characteristics that should cover the water and how much water consumed with full sample working
HI FRIEND I CAN HELP YOU !! you must use a desionized filter thats could be a cartriged with a osmosis reversed unit, this equipmet uses a internal tank because it has a internal compresor.
400 tests/hour, or
560 tests/hour with ISE uni
Sample volume
2μl-45μl, with increment of 0.1μl
À„ Water consumption
20L/hour because it washes before all cuvettes and after too.
Water Supply and Drainage Requirements
À„ Water: The water must meet requirements of the CAP Type II water.
À„ Water pressure: 100~392kPa, with peak no greater than 800kPa.
À„ Water flow: Continuous flow no less than 30L/hour, with peak no greater than
1L/minute. An internal water container (above 40L) is required for the water unit
which has a flow of 20L/hour.
À„ Water temperature: 5-32.
3 Installation Procedures 3-3
À„ The water used on the system is supplied by a wash unit or other pressure water
supply. The tubing between the inlet filter and the water unit(other pressure
water supply) must not exceed 8 meters.
À„ Water supply module:
a. The tubing between the inlet filter and OUTLET of the water supply module
b. The tubing between the INLET port of water supply module and cover of the
deionized water tank must not exceed 10 meters.
c. The volume o the deionized water tank is 80-120L.
d. The tubing from VENT of the water supply module to the sewer must not
exceed 10m.
e. The water supply module should be placed on the ground.
À„ The inlet filter should be placed on the ground.
À„ The tubing from the outlet on the system to the inlet filter must not exceed 2
meters.
À„ The system drains away the low-concentration waste directly.
À„ The tubing from the outlet on the system to the waste sewer must not exceed 5
meters.
À„ The waste sewer must be higher than the ground within 100mm.
À„ Drainage module:
a. A drainage module is required if the environmental conditions for waste
discharge are not met.
b. The tubing between the outlet on analyzer and inlet of drainage module
must not exceed 5 meters.
c. The tubing between the outlet of drainage module and the sewer must not
exceed 10 meters.
d. The sewer must be within 1200mm higher than the ground.
e. The drainage module should be placed on the ground.
À„ The high-concentration waste bucket should be placed on the ground. The
tubing between the outlet on analyzer and the high-concentration waste bucket
must not exceed 2 meters.
i hope you help