OEC 9600 buzzing alarm

9600 c-arm intermittently buzzes and will not fluoro from foot pedal (but will from control on c-arm). Any ideas? Rebooted, worked fine. But happens randomly multiple times a month.

Renee@mediserv-online.com


   RE: Mobile C-Arm Report Post
Wed Apr 23 2014Reply from xrayxpert
avatar placemark

Have you figured this out yet? If not....are you getting any errors? What type of sound and where is it coming from? Does the fluoro lamp light up upon problem? Have you checked the footswitch and cables?

What all have you tried?

 


Report Post
Thu Apr 24 2014Reply from MES, Inc.
avatar placemark

Thank you for your inquiry.  The problem was corrected by replacing the analog interface board.  The foot switch and cables were already changed out and did not correct the problem.  The buzzing was actually a beeping noise.  We weren't able to make exposure with the foot switch when this happened, but the exposure button on the console worked.


Report Post
Thu Apr 24 2014Reply from Cbet_Matt
Cbet_Matt

Thanks for letting us know how it was resolved. That will help others to shorten their down time. That goes for me too.



Matt

 